I remember the chiralpak AD only allow the ratio of alkane/ethanol to be 100/0 - 85/15 or 40/60 - 0/100. The ratio outside of this range might have some adverse effect against the column. Is it safe to use at 70:30? Is there any other list of separation enhancer we could use such as Diethylamine?
Thanks,
According to the manufacturer, the baseline stability may be adversely affected, presumably not the column. In my limited experience with a Chiralpak OD-H column, ethanol can provide improved resolution, but the lower ethanol content can cause standard pump seala on an Agilent 1100 to leak after a few hours. Polyethylene normal phase seals have to be used.
Typically, the potential use of TFA or DEA depends on the functional groups on the enantiomers you're trying to separate.
http://www.daicel.co.jp/chiral/e/produc ... on/ad.html
" Some of the most popular eluents for HPLC (such as acetone, acetonitrile, chloroform, DMF, DMSO, ethyl acetate, water, methanol, methylene chloride, and THF) remaining in your HPLC system may destroy this column....Suitable Mobile Phases include:
- Hexane/2-Propanol (100/0 to 0/100 v/v) Typical Mobile Phase used is Hexane/2-Propanol (90/10 v/v).
- Hexane/Ethanol (100/0 to 85/15 v/v), (40/60 to 0/100 v/v)
(NOTE: The range 85/15 to 40/60 (v/v) may adversely affect baseline stability.)
- Suitable Mobile Phase Modifiers include N,N-Diethylamine for a basic sample, and Trifluoroacetic acid for an acidic sample.
(NOTE: Minimize use of the modifier. Typical use is 0.1%; maximum 0.5%).
Bruce Hamilton