Intermittent Erratic Baseline
Posted: Wed Jul 12, 2017 2:49 pm
Hello everyone,
I'm new around here, and fairly new to GC/MS, so please call me out for any omissions or oversights. The system experiencing an issue is an Agilent 7890A GC, 5975C MSD in positive CI mode, helium, currently measuring glucose pentaacetate. Been running this protocol and method for many years.
A few months ago, I started observing small, infrequent disturbances in the baseline, sudden rises and falls that look like steps. (see image below). The issue has gotten worse, but frustratingly intermittent. I can run 100 samples with no problem then suddenly the noise will become so severe that the data is unusable, then a few samples later the problem is gone. (see image below) The baseline issue happens even when I run air and methanol blanks. I disconnected the column from the MSD, and the problem still occurs, but when I switch out the CI source for an EI, the issue seems to go away. At first I thought it was the electron multiplier, but we're tuning at ~1600 V. Another factor: our filaments have started to burn out after two weeks, and they used to last for months. Tune shows no obvious sign of a leak (autotune is fine, no peak at 32) but any advice for doing a more thorough check would be appreciated.
There is black build-up on the source body near the filament that we can't completely clean off. (see image below) Also, the repeller insulator looks dirty, but everyone is too afraid to clean it if we're not sure it's the issue (those things are not cheap..) I also noticed black buildup in the metal tube that holds the source in front of the quadrupole, near where the filament is. I'm not sure if this is a result or contribution to the issue, but it has looked this way for a while.
I've checked all the connections, system is plugged into the same outlet it always has been, no new equipment nearby. Two other GC/MSs (one CI one EI) working fine. Replaced all the usual consumables on the GC side, swapped out the CI interface tip seal.
Any idea what's going on?