1) It is difficult to inject water onto GC columns (but this is not at all impossible).
2) the higher resolution of GC is not really necessary in pharma, because a typical formulation contains a limited number of ingredients (compared with something like an environmental analysis).
3) The quantitation aspects of capillary GC are seen as more suspect than HPLC. In split and splitless injection at least the whole sample is not directly introduced onto the column as in HPLC.
4) Many problems with polar compounds on GC are more historical and relate to use of more active packed columns. The pharma world has just not moved to embrace the newer techniques.
5) The flexibility of separations in HPLC is much greater due to the ability to change selectivity with changing the mobile phase- but often, pharma methods are set up and never touched anyway because of validation requirements.
6) Perhaps I got it all wrong and a much larger proportion of drugs is involatile than I had thought.