EZchrome-Creating Calibration curve

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hello,

I am trying to create a calibration curve for one analyte at a number of different

concentrations level.

I managed only analyze one injection every time and I receive each time only

one point instead of all seven injection in different concentration level.

Can you please send me the right proc ? Which parameters are critical to

create the cure?

Thanks!
in order to make a calibration curve you need to enter the concentrations in the compound table of the methods under the Levels. Each level its unique concentration. For 7 levels you fill out all 7. You can go up to 99 levels if you like to calibrate :).
Then in the sequence you tell which level corresponds to which calibration run and tada the method is calibrated when you reprocess.
Freek Varossieau
OpenLab CDS 2 specialist
BeyondOpenLab
beyondopenlab@gmail.com
+5977114721
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry