Retention time shift / Peak height drop
Posted: Thu Jul 06, 2017 5:33 am
I am running the Vitamin D3 method by USP on an L8 aminopropylsilane column. The peak retention time of the Vitamin D3 should be around ~4.7 minutes.
Mobile phase A 99%: Hexane
Mobile Phase B 1%: IPA.
Flow 1.0 ml/min.
After running the D3 method numerous times and having some consumables changed from time to time, I have seen the retention time shift slightly which is not that unusual to see after continuous usage of the same column.
Recently while I was out from work, our service engineer was called in due to degasser fault. The degasser chambers were flooded and were replaced. The chemists in the lab tried running another sample of Vitamin D3 but the pressure was too high. The service engineer found that the flow cell was the cause of the high pressure so he rebuilt the flow cell. The engineer also replaced the gradient proportioning valve.
The chemist ran a Vitamin D3 standard after the above items were replaced and found that the standard had decreased to half of the normal peak height. Typical peak height for a 3.0 ug/ml standard solution is around ~110,000. We are now getting around ~50,000. Peak retention time for Vitamin D3 has more than doubled what it used to be (Vitamin D3 has been roughly around ~5 min).
The chemist thought that the column was the issue so another brand new column was ordered. The retention time on the new column was around ~11min.
We only run one other test on our HPLC right now. We analyze CoQ10 and the retention time for this has been coming out fine. I initially thought the issue with the Vitamin D3 could be due to the gradient proportioning valve but since the engineer changed that and the CoQ10 mobile phases are mixing well that can be ruled out.
I have been trying to troubleshoot this by doing the following:
1. Checked the delivery volume of the mobile phase reservoirs. They are fine.
2. Checked the system for leaks. Pressure is at the usual ~540 psi (although I know there could still possibly be a leak somewhere even if the pressure is steady)
3. Washing the column with the recommended mfg instructions.
Any other recommendations? I want to get the retention time back to the established time.
Flow 1.0 ml/min.