lmh » Fri Jun 30, 2017 9:15 am
In this case I wouldn't bother doing a formal scaling.
I assume from the change of column that the OP needs high resolution and doesn't care about run-length. They haven't given the particle size of the new column, but switching to big particles in a long column is typical for high-resolution methods(*).
If the aim is to get the highest resolution, obviously we want the lowest point on the van Deemter curve, and we're balancing band-broadening effect of general diffusion (worse at low flow) against the band-broadening effect of diffusion in and out of pores (worse at high-flow). Kinetex is a solid-core column, so diffusion in and out of pores is reduced, meaning that high-flow is generally good. 4.6mm Kinetex columns aren't rated to high pressure (2mm columns are), so it's likely the limiting factor will be pressure; I'd go for a fairly high flow, maybe 1mL/min, and see what it looks like...
There's certainly little point in scaling directly between these two columns, as you wouldn't make that column change unless you wanted to change the performance of your chromatography, and you may as well take advantage of the change to optimise for what you need.
(* small particles give short run-time, not high resolution. Halving the particle size doubles N but quadruples pressure, so if you're limited by the pressure your system can handle, to maintain constant (maximum) pressure, you need to reduce the column length 4-fold, so you halve N overall; the advantage is that the method gets 4-times quicker at a cost of only root-2 loss of resolution).