Hello. I have a HPLC HP 1050, with autosampler 1100.

Anybody know how to calibrate pump delivery volumen vs method set volumen?
i meassured the delivered volumen and it is less than my method.

It is possible to fix this problem with a calibration, or i must to change any component/s?

Thank you very much.
Regards.
If set volume and delivered volume do not match I would first suspect issues with a check valve or possibly a leaking piston seal.. I would at least attempt all things related to physical issues before even thinking about adjusting anything.. if it is even possible..
Kind regards
Leadazide
By what percentage is the difference?

In general, read what leadazide posted. I'd check for leaks first at the outlet of the pump module, and then add the other modules one by one.

One often-overlooked place is leakage out the purge valve, which oftentimes has waste tubing attached to it.
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 3 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot], Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry