Agilent 1260 ELSD and Empower3
Posted: Fri Jun 30, 2017 6:16 am
by heiko
Hello, we want to retire OpenLab, but waters guy in our country told us that Agilent A1260 ELSD cannot be contolled by Empower3. Is that true? If not, which ICF is available?
Thanks in advance.
Re: Agilent 1260 ELSD and Empower3
Posted: Sun Jul 02, 2017 8:27 pm
by DR
My guess is that it is just a matter of time before ICOP is available for that detector.
In the mean time, (assuming you have another detector in line w/ the ELSD), you could use an eSATIN to collect ELSD data and just leave that module out of the Empower definition for that system... (not tried personally, but will probably work). Duplicate SSM using the eSATIN's method set.