Total Publisher
Posted: Fri Sep 15, 2006 11:00 am
by sadsal123
Does anyone have a total publisher manual?
I am trying to create a report on it but cant find the %Amount column that we create in the Total Chrom report. Can anyone please help?
Salma
Posted: Fri Sep 15, 2006 11:06 am
by sadsal123
Sorry! That was meant to be TC Publisher
Amount in Publisher
Posted: Tue Sep 19, 2006 9:19 am
by kdamme
As you might have noticed, TC Publisher is a complex utility. As far as I know there is no manual, but the Help-function is not too bad.
To add an amount column:
select either the standards or the non-standards box. (A report is valid for either a standard or a sample, unless you make a report valid for both..)
Right click and modify your properties. Click columns --> Add --> Peak properties --> AdjustedAmount (doubleclick)
This should give a your amount in the Publisher Table.
If this will not work, please let me know.
Kevin
Posted: Tue Sep 19, 2006 11:56 am
by sadsal123
KDAMME, Thank you for your reply. I might not have been very clear in my query and I apologise for that, the Problem I have is that after I loaded the data of a result, the amount(%) that I have in Total Chrom Workstation Report is not the same that I got on TC Publisher (in the AjustedAmount Column).
As for manual, Perkin Elmer very kindly sent me a user guide document yesterday. So if you every require it, please feel free to ask
Salma