Noise after a post-column reactor
Posted: Thu Sep 14, 2006 4:25 pm
OK, I've posted a few times already trying to figure out why my peaks wouldn't resolve they way they would on a second system. I'm happy to report that I have successfully traced the problem to a back-pressure regulator, but as a result I have a new problem.
I have to place the regulator between the post-column reactor and the detector. If I don't, my baseline is poor (a lot of heavy noise, the up-and-down oscillating pump-noise pattern type) and I can't integrate well. If I do, my baseline is nice and flat but my system efficiency is gone (peaks merge in with each other).
I do have a 100psi BPR after the detector. A thought I had was to try lower and higher regulators to see if they have any effect.
When I had a very short connection between the reactor and the detector I also had a lot of noise, that's why I added the BPR. I have noticed that the stream coming out of my reactor is bubbling, which makes sense since the temperature is >120°C and venting to atmosphere. So I concluded that with the short connection reaction bubbles were getting into the detector. On my other system the problem seems to be solved by having extra tubing (about 1.2m) between the reactor and the detector in order to give the system time to cool (just speculating). I have since added a similar about of dead tubing between the reactor and detector.
My detector is actually two self-contained UV-VIS preset to 570nm and 440nm (yes, amino acid detection with ninhydrin). I have also discovered by switching the order they occur in the stream will change the noise found in each detector (second one has less noise). As a result I'm pretty sure something needs to be in the system to bring down the noise.
I'm a bit hesitant to take this to Pickering since I'm not currently using one of their products, but I'm pretty sure they would certainly be a good souce for information. If anyone has any ideas on how I can bring down my noise without sacrificing my efficiency I'd appreciate it.
Thanks in advance
Thanks in advance