I am in the process of analyzing the composition of unknown samples that contain H2, O2, N2, CO, CH4, and CO2 on a capillary column (Carboxen 1010, oven at 25C for 10 min then 5C/min to 60C, Ar carrier). The samples are collected in miniature cylinders during experiments, with varying pressures, which I can hook up to fill a 10 µL evacuated loop on a six-port valve, while reading the pressure in the loop on a gauge, to know how much is really being injected.
What is the best way of performing a calibration to determine a % composition? I have a Scotty calibration mix with 4-5% of all gases above in helium, so all that I can do is vary the pressure injected and plot partial pressure of each gas vs. peak area. I did this and got some very nice, slightly nonlinear curves, however when I then use these curves to get partial pressures in my unknown, the total pressure is far less than what I read on the gauge! It is possible, but unlikely that this is really true. Is there a better way of doing this?
Several months ago, I did some very detailed calibration curves of %H2 and %air vs. peak area for varying compositions of 44.7 psia of air/hydrogen mixtures, which I can make myself, but this curve (a line with R^2 = 0.997) predicts a zero peak area for 22% hydrogen! And this was actually the case when I ran a sample with 22% H2. This, by the way, is in contradiction to the partial pressure vs. area calibration curves generated from the calibration gas.
Basically, your system seems fine ( assuming that you are evacuating all the way back to the cylinder valve to clear the line, rather than just purging, and you are allowing an equilibrium to form in the loop.
Such a small loop implies the samples are at high pressure, in which case you should match the pressure with your standard mixture. I'd actually use a larger sample loop ( 100 - 500 ul ) for near-ambient pressures, but it depends on your system, and loop pressure ).
The problem appears to be your calibrations, and possibly the fact that your standard mixture contains helium. I assume your system is leak-free, and that you are purging the sample lines.
Provided the sample doesn't change, you may find using a constant sampling pressure will give more consistent results. I would definately want at least one standard mixture concentration and composition that more closely reflects your sample composition and pressure.
If you know the range of your samples, it's good to bracket them. You can always use dried ambient air for the O2 and N2, taking care of where the argon could appear on your column, which I'm not familiar with.
The alternative is to use different-sized sample loops, which if you have, I would strongly recommend using to investigate your problem and the linearity versus pressure for each compound calibration.
The issue with the hydrogen may simply be the well-known behaviour of thermal conductivity detectors and typical GC carrier gases with varying hydrogen concentrations in samples. The hydrogen thermal conductivity is so different that as the concentration increases the conductivity passes through the baseline. There are carrier gas blends that overcome the problem.
Please keep having fun,
Bruce Hamilton