So I have a 7820 that I am trying to work on and it is giving me fits. First off, it is a p/p and s/sl with dual FID. There was not an ALS installed and I installed that. Before installing the ALS board, everything worked like it was supposed to. Detector flows as shown on the GC were the same as measured out the detector and all pressures on the inlets were correct. Now, after the ALS has been installed, the front FID board takes a nose dive and blows something and when you turn on the GC, it will not pass initial self tests. When I remove that board, everything is fine, except I cannot use the front detector. Second, when measuring flows on the back detector, lower flows (such as 40 mL/min) cannot be achieved according to the GC but when measured out the detector, I am reading upwards of 140 mL/min. If I increase the flow to 100 mL/min, the GC thinks it is reading that, but measured flow is 15 mL/min. Last, I cannot maintain pressure on my split/split-less inlet. It reads a constant -0.5psi no matter what I set the pressure to.
I have completely leak checked everything (no leaks.) I initially had the plumbing routed to both inlets and detectors but have only connected to one inlet/detector and still have the same issues. Both inlets have new liners/septum/gold seal (where applicable.) I have also performed a jet test on the back inlet and H2 set to 75mL/min reads 0.22 mL/min on the makeup.
I have also zeroed the channels according to the manual for all involved. I also have updated the frit details for each FID as downloaded from the latest file from Agilent.
If anyone has any suggestions, please let me know. Thank you.