Molly » Sun Sep 10, 2006 3:47 pm
Hi, Sergio:
Here is just my opinion on your method. Monolithic column is good for very fast separations (1-2 min) at high flow rate (5ml/min) and for relative simple separations. The peak shape is somewhat worse than typical packed columns and I found the narrow bore monolithic column tends to have fronting in peak shape. If you want to have more plates/peak capacity, I would suggest you try a typical C18 column (like Agilent SB-C18, Waters Symmetryshield RP18, or Luna C18) at 50X4.6 (3um) at 2ml/min and run the same 3min gradient. If you can modify your 1100 to do high throughput separation(you can buy kit from agilent so that extra column volume is minimized), then the new sub 2um column is a good choice for fast screening as well. If your method does not need to be MS compatible, 0.1% H3PO4 will provide better baseline and sensitivity at low detection wavelength. For MS friendly method, I saw more people use TFA than formic acid, but I do not know why. The re-equ volume has to be at least 1 column volume+ twice dwell volume, so 0.5min proposed might not be long enough. For a typical 5cm column, void volume is 0.5ml, and for Agilent 1100 with quaternary pump, dwell colume is 1.1mL (smaller for binary pump), so the total re-equ volume need to be 0.5+2X 1.1=2.7mL, at 2ml/min, you need to flush at lease 1.3min to fully equlibrate your column.
Thanks, Molly