We are pharmaceutical company (new) that starts preparing to FDA inspections.
We have consulted regarding certain HPLC topics and I wanted to ask if they are usually performed in FDA regulated pharmaceutical companies:
1) We perform system suitability test at the beginning of the work (system suitability solution, control standard preparation, standard repeatability (5 or 6 injections) and then standard every 6 samples or after 3 hours and the end. We were consulted that FDA requires performing all the initial system suitability tests (including 5 standard injections) also at the end of work and in the middle of the work, if it takes more than 8 hours.
2) Empower 2 is not FDA compliant, and we have to switch at least to Empower 3.
3) When we switch to Empower 3, we would have to revalidate all the methods.
Please advise, if FDA regulated lab should act according to all this requirements?
If revalidation is really required, is it full revalidation? Including stability indicating and all?
Thank you very much!