high pressure problem!
Posted: Tue Aug 29, 2006 3:36 pm
Sorry if asked before!
I have luna C18(2) column 150 x 2mm. Because I share the LC system with other people, So I have to took off the column (store in 100% ACN) after my run. The problem is that everytime I put my column on after couple days (other people use the system), I got high pressure!
My mobile pahse: 0.1% acetic acid; ACN.
But if leave the column on and next day I have no problem to run it. I do wash the system (other people use MeH and water without buffer) before I install the column, but still no good!
So, I have to wash my column quite long time (at least 3 h) at 0.1 ml/min before I convert to 0.2 ml/min. As long as it runs the retention time is good and peak is sharp. What is wrong?
I have luna C18(2) column 150 x 2mm. Because I share the LC system with other people, So I have to took off the column (store in 100% ACN) after my run. The problem is that everytime I put my column on after couple days (other people use the system), I got high pressure!
My mobile pahse: 0.1% acetic acid; ACN.
But if leave the column on and next day I have no problem to run it. I do wash the system (other people use MeH and water without buffer) before I install the column, but still no good!
So, I have to wash my column quite long time (at least 3 h) at 0.1 ml/min before I convert to 0.2 ml/min. As long as it runs the retention time is good and peak is sharp. What is wrong?