HP5973..autotune problem
Posted: Mon Aug 28, 2006 3:42 pm
by Rick
Hi!
I have clean my source from my HP5973N but repeller give a high value (34.

and entrance lens is low (9)...normally, repeller give 24 and entrance lens give 16...i have reclean 2x my source but i have again the same value.
Is it possible that my electron multiplier horn is the source of my problem?
Posted: Wed Aug 30, 2006 10:14 am
by chhubert
Rick,
Did you reset the autotune value to default prior to tuning? If autotune is started with previous autotune settings, the tuning algorithm simply re-adjusts the settings back and forth a little to achieve your pre-set criteria. Try and see the difference if you start tuning after resetting the tune values to default first.
Posted: Thu Sep 07, 2006 6:28 am
by jimmyhu
I would like to share some interesting findings about Repeller figures on tune report I observed on my Agilent 5973N. The Repeller figure was 32 when a tune was performed using a method method (Flow rate: 0.5 mL/min) and the Repeller figure was 25 when a tune was performed using a normal working method (Flow rate: 1.0 mL/min).
Jimmy
Posted: Thu Sep 07, 2006 6:32 am
by jimmyhu
Sorry about my last post that contained a typo. The method at flow rate of 0.5 mL/min means standby method.
Jimmy