pressure fluctuation

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
Hi,

I'm having issues with the pressure fluctuating/not stabilizing in my Agilent GC/MS 6890. The EPC was replaced and fixed the problem for one night of sample analysis. Agilent techs helped with troubleshooting and turning the gas saver off is what ultimately solved the issue, however now it just sits on the preprun and doesn't register the parameters for the method. Any advice would be appreciated
"doesn't register the parameters for the method" Doesn't register where? For the time being, take the workstation out of the equation. If you manual set values on the GC, is the pressure stable? Going to need a bit more information.
Ok well the problem is solved after changing the liner and septum, however, stopped again after two samples ran. And no it doesn't stabilize if the values are manually set
Is your gas supply stable? Regulated?
Check for leaks, using a leakseeker, not the onboard diagnostics.

Peter
Peter Apps
yes gas supply is stable and uses a regulator. I think I'll put a call in to Agilent and troubleshoot it again.
6 posts Page 1 of 1

Who is online

In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry