- Posts: 4
- Joined: Tue Jun 05, 2012 3:16 pm
I am using Agilent 7890 GC and 5975C MS, with a HP-5MS column. The inlet was set for 220 degree with splitless mode. The purge flow is 50 mL/min and the carrier gas helium is at 1 ml/min.
The method runs very well in the past until recently I had two manual injections since I had very limited sample volume. All of a sudden, this huge solvent peak appears and masks everything.And even I can see my analyte at a much alter time, the sensitivity was reduced a lot.
I thought it might be the inlet problem and I have opened the insulation cap, cleaned the gold seal, replace the septum, the liner, the o ring, cut my column twice,each time around 15 cm, re-installed the column many times.................The pressure at the inlet reads OK as normal at 7 psi at that flow rate at 220 degree. How can I check anything else?
The peak is still there, even if I inject nothing, the residue from the wash station of hexane produces a big peak at 8 minute.I think this should never happen.
Should I replace my column?