Posted: Mon Aug 21, 2006 2:02 am
Can anyone help me figure out the way of analysing gas compound from airbag and quantification based on the built calibrations done by liquid compound injection? Or I have to prepare gas standard in a same airbag like the sample?
The situation is that I made toluene standard in CS2, and made 1 uL injection. It is 322 uL when it is injected in jection port using HP Pressure/Flow Calculator. A calibration curve can be obtained for toluene with the knowm amount of toluene and the integrated area obtained from GC run. Then I injected 500 uL of gas sample from an airbag in which the toluene is present.
If the quantification of the gas sample based on the above calibration curve is given to be 100ug, so a factor of 0.644 from 322uL/500uL is used to correct the raw data of 100ug. The final result should be 64.4 ug (100 x 0.644). Is it correct?
Cheers
Jimmy
The situation is that I made toluene standard in CS2, and made 1 uL injection. It is 322 uL when it is injected in jection port using HP Pressure/Flow Calculator. A calibration curve can be obtained for toluene with the knowm amount of toluene and the integrated area obtained from GC run. Then I injected 500 uL of gas sample from an airbag in which the toluene is present.
If the quantification of the gas sample based on the above calibration curve is given to be 100ug, so a factor of 0.644 from 322uL/500uL is used to correct the raw data of 100ug. The final result should be 64.4 ug (100 x 0.644). Is it correct?
Cheers
Jimmy