If i'm correct you and another lab are getting different values for the same samples right?
Getting the same results on the same samples with a different instrument, a different injection, and a different person performing the analysis can be troublesome. What errors are we talking about?
Thanks for reply. All of you are saying is true but the error is two folds, i mean if with HS injection i get 400ppm then for Direct Injection the value is 800ppm. So that is why i think the problem should be something else and not a common mistake.
Hello Lucas
I we are to help you we need an accurate description of what the problem is; in your first post you say "The problem is that both EtOH and Npropanol gives higher values of concentration using HeadSpace" which is exactly opposite to your most recent post.
We need to know what intruments you are using, your analytical conditions and how you are doing your calibrations, and the same for the other lab.
Have either of the labs ever analysed a refererence material with a known concentration ? What were the results ?
Peter
Hi Peter. Thanks for your reply. I agree with you, i missed up. Let me be more specific. My lab recive a group of samples witch contain MeOH, EtOH, and Npropanol in brine as solvent. Since we start our values of concentration (for EtOH and Npropanol) are twofold than the other lab (They are a private lab so they dont want to share anything). The opposite happens to the MeOH, sometimes values are same but othertimes our values are lesser.
Both Labs must report results to the same company.
We use a Agilent GC-FID with HeadSpace Injector. Every year we do the offical service and maintenance for all the equipment.
For calibrations the company provides the brine, so we just prepare differents concentrations (100, 200, 400, 600, 800, 1000ppm for example) for triplicate each. Then we make the numerical regretion and start to measure. We do all at the same day. Also we prepare checkpoint samples (put it random on the sequence) to be sure that all is running well.
On the other hand the samples are not coming direct to us. First the company send the samples to the private lab and then they send it to us. The samples comes in a plastic 20ml bottle. Then we take an aliquot of 2ml and put it in a HS-glass vial sealed with a metalic cap that contain a rubber septum.
We asked they to measure a reference material but they answer that is not in their contract so they wont do it. All I know about their equipment is what I posted first.
Think that is all. Thank you for your support.