Problems with the pressure

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Dear members of this forum,

I have problems with the solvents in the HPLC where I work. I am very strict with the variation of pressure when I am runing, the mobile phase is 100 % acetonitrile.
The variation of pressure is about 100 psi and tne average is 1000 psi. The technical service told me the variation is about to 40 psi not more. The acetonitrile is degased but I am convinced the problem is the degasification in line but I do not know.
I hope this problem happened in anybody of the forum. The model of HPLC is Waters 2690.
Thanks in advance for your help,

Diego

If your suspicion is the degasser, you should try it without the degasser.

How looks the spurt out of the capillary to the column by a flow speed of 4 mL/min?[/quote]

I guess that you have an air bubble in one of your check valves, or in the pump head. Try to purge the pump a couple times, otherwise ultrasonate the cheackvalves
3 posts Page 1 of 1

Who is online

In total there are 34 users online :: 3 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry