ym3142 » Thu Aug 31, 2006 4:08 am
please clarify: 8/30 you said
The fronting is experienced when the injection volume is doubled.
8/28:
The peak shape of my std injection is fine when the injection volume is 50 uL, but terrible fronting (almost as if another peak is emerging in front of the peak of interest) is observed when the injection volume is reduced to 25 uL.
which is your case?
q: when you changed the volume, did you change the concentration?
if it was former, you overloaded.
if the latter, it does not make sense if the concentration was the same. so if the latter, can you go back repeat this like concecutively inject the same solution from the same vial 50ul, 25 ul, 50 ul and 25 ul to see repeatability .