priz » Thu Aug 17, 2006 4:02 pm
Dear Danko,
Thanks so much for your suggestion. I've actually been trying what you suggested.
I've set my excitation at 700nm and my emission at zero-order to detect emission at all wavelengths. I find if I do this, I do detect chemiluminescence peaks. I've also been experimenting with physically blocking the excitation path at the flow cell. The effect of blocking light gets a little tricky to interpret with the calculations that Chem Stations does with the excitation and reference diode.
Right now, I am able to detect my chemiluminescence peak, but my working standard curve is 8000 - 30000 pmoles when my reference paper is detecting 10-600 pmoles. My background baseline is high (500 luminescence units) even when only my mobile phase pump (Methanol:tert-butanol 9:1) is on and not my chemiluminescence reagent pump. I'm concerned about the high background baseline. I guess I'm worried about swamping the detector (I come from a GC-FID background) if I'm starting with such a high baseline.
The noise is very high, but I have my photomultiplier tube (PMT) gain set high, 15, on a scale of 0-18 where 10-15 is used for "normal" applications according to Agilent. I think I need a high PMT gain for better signal to noise. If I reduce my PMT, I lose my peak. I'm going to work on finding a better balance. In hopes of reducing noise, I'm going to treat my solvents with a chelex resin or molecular sieves. I also plan to increase the response time to lower noise. Today I will try increasing the response time and adjusting the PMT.
If you have any other suggestions or ideas or know of anyone else who is doing this, please continue to share with me. I'm still not sure if my system is actually going to work for my application.
Danko, thanks again.
Priz