Rob, I think setting a fairly tight spec for fronting is a good idea if you have found that fronting provides warning of impending problems. System suitability "suggestions" from FDA/ICH/USP are minimum suggestions. It's up to us as chromatographers to make them as tight as necessary to ensure validity of results.
My take on the situation is that fronting is fairly rare and usually associated with inadequate buffering, solubility problems in the mobile phase, tendency of analytes to form aggregates, or protein conformational changes (partial reversible denaturation). Most of those can be fixed.
Hans and Peter are both right in their comments on LOD - a broad peak will have a higher LOD than a sharp one. However, there are many cases where LOD is not an issue (pharmaceutical potency or content uniformity, for example, where you only have to demonstrate linearity over a range bracketing the nominal concentration). As long as the calibration plot is linear (i.e., only peak amplitude changes with load) then the peak shape is of secondary concern. That said, I like to think of myself as a craftsman in chromatography

, and skewed or squashed peaks offend my sense of aesthetics.
Danko, I think your triangle test can be useful, but is arbitrary, because the numbers depend entirely on how you do your normalization (I can make the number 1000 times bigger by simply setting 1 μV = 1 sec instead of 1 mV = 1 sec). Having said that, if
you pick an appropriate normalization, I can see it being a useful system suitability criterion. I'd hate to have to justify it to an auditor, though.