- Posts: 52
- Joined: Sat Jun 27, 2015 12:03 am
I'm having a problem with the alcohols not getting to the GC. I'm running a 5890 S II with a 5972 MSD an Archon Sampler and OI 4560 P&T. We had been having Alcohol resolution problems for a while but didn't think much of it as the instrument was due for maintenance. Then they just disappeared. We're running our waters in soil mode for a variety of unrelated reasons. the first thing i did to check was direct inject 1 uL of our working EtOH TBA standard and sure enough right where they're supposed to be. So far i've done the following maintenance in this order:
1: inlet new liner septa gold seal and column trim 2: New Trap (#10) 3: new transfer lines (sulfinert) 4: new water management unit and assay line 5: new soil purge needle and pencil filters on Archon 6: sonicated the sparge mount and 6-port valve in MeOH replaced the line from the sparge mount to the 6-port valve new sparge tube replaced all valves on the manifold.
After all this i still can't see the alcohols, not even a "might be it" blip. we run the P&T at standard conditions with a 0.5 minute desorb 11 minute purge at 40 C. At this point i have no idea what else to even look at. does anyone have any thoughts?