I am trying to analyse beerepellents
(Benzaldehyde, Phenol, p-Dichlorobenzene, Nitrobenzene, p-Bromochlorobenzene(ISTD), Naphthalene, Thymol)
in honey using simple Headspace
(5 g honey, 2 g NaCl, 2.5 mL H2O, 20 mL Headspace Vial with screwcap)
Incubation at 80°C for 2 h
Injection 1000 µL
Split 1:5
Injector: 250 °C
Carrier Gas: He
flow rate 1 mL/min
calibrations are perfectly linear
however depending on the honey type (forest, polyfloral, limetree,..)
the calibrations give totaly different slopes
Only for Phenol, p-DCB and naphthalene the calibration is independent on the honey type used.
Any Ideas how to solve this problem or what could be the cause for this?