Just curious what do your think about the most challenging/most difficult part facing the LC/MS analysis? How do you usually deal with it? Thanks.

Posted: Sat Aug 12, 2006 5:51 pm
by Uwe Neue
Depends on what you are doing...

If it is analyses from a biological matrix, I would think that ion-suppresion is the general problem.

Posted: Sat Aug 12, 2006 11:06 pm
by meillid
Uwe, Thanks for your answer. What do you do to minimize the ion suppression effect? What else beside ion suppression? Thanks so much.

Posted: Sun Aug 13, 2006 4:43 pm
by Uwe Neue
To reduce ion suppression stemming from a complex matrix such as biofluids, you need to reduce the complexity of the sample with solid sample clean-up techniques such as SPE and high-quality chromatographic separations.

You can test for ion-suppression issues by doing post--column infusion of the analyte(s).