First off, the setup:
GC Model: Agilent 7890B
Columns: HP-Plot Q 0.53 mm x 15 m, CP-Molsieve 0.53 mm x 25 m
Detectors: TCD-Methanizer-FID
Carrier: Argon (30 ml/min)
We are using this setup to detect trace hydrocarbons from the electrochemical reduction of CO2 in the gas phase. Products concentration is very low so the main component in the sample is CO2. Products we detect are CO and H2 (mainly) and then light hydrocarbons (methane, ethane, ethylene). With some valve switching, we can exhaust all the CO2 before the methanizer and with the current method, we can resolve all peaks quite well. The method takes 30 min roughly (this is without any oven ramp).
The issue I face is that every time I ramp oven temperatures with the FID in-line, the FID flame extinguishes. This also happens with increased column flows so I'm guessing it may be a column pressure issue? The FID baseline is quite low, ~0.5-1 pA when the TCD and FID are in-line. ~1.5 pA when the FID is isolated. It has been like this since the installation (about a month ago). All gases are UHP from cylinders with H2 flow = 45 ml/min and Air flow = 500 ml/min (Agilent recommended for methanizer-FID).
Any suggestions on how I can stabilize the flame?
Cheers