High molecular weight compound analysis using GC-ECD
Discussions about GC and other "gas phase" separation techniques.
7 posts
Page 1 of 1
We have one fluro-substituted compound. The molecular weight of the compound comes around 1000. Is it possible to analyze the compound in GC-ECD? What is the maximum molecular weight we can analyze in GC? Thanks in advance your help.
-
- Posts: 404
- Joined: Wed Nov 09, 2016 12:43 am
There is no MW that cannot be analyze by GC. However, GC requires that the molecule be volatile which decreases as the MW increases. You are more likely to analyze the molecule with HPLC if it can be solubilized (I analyzed the peptide Oxytocin which had a MW of 1007 daltons).
-
- Posts: 4
- Joined: Thu May 11, 2017 10:57 pm
Thanks for your response. I am also thinking of moving to UPLC. Some articles are saying you can analyze some alcohols by derivatizing it with Heptafluro derivatives. The derivatizing process going to increase the molecular weight. I was wondering how high we can in terms of molecular weight we can go. Most probably I will analyze them using UPLC.
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
Derivatization of alcohols and acids for GC (for instance esterification) will indeed increase their molecular weight but increases their volatility. It lowers the intermolecular forces. There is no easy relationship between MW and boiling point unless you're talking about a series of linear alkanes.Thanks for your response. I am also thinking of moving to UPLC. Some articles are saying you can analyze some alcohols by derivatizing it with Heptafluro derivatives. The derivatizing process going to increase the molecular weight. I was wondering how high we can in terms of molecular weight we can go. Most probably I will analyze them using UPLC.
There's no simple rule whether something is more suitable for GC or HPLC, or both. Check literature or application notes how others do the analysis of yours or similar compounds.
-
- Posts: 4
- Joined: Thu May 11, 2017 10:57 pm
Thanks for your response. I will do some literature search on this analysis.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Unless it is perfluorinated an ECD (as in electron capture) will not give much more signal than an FID. With a MW of 1000, even with no sticky functional groups you ate not going to be doing this with the usual columns and conditions.We have one fluro-substituted compound. The molecular weight of the compound comes around 1000. Is it possible to analyze the compound in GC-ECD? What is the maximum molecular weight we can analyze in GC? Thanks in advance your help.
Peter
Peter Apps
-
- Posts: 4
- Joined: Thu May 11, 2017 10:57 pm
Thanks for your response.
7 posts
Page 1 of 1
Who is online
In total there are 22 users online :: 3 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Bing [Bot] and 19 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Bing [Bot] and 19 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science