I have problem connecting the G1315B DAD with chemstation. It says the
detector s not propely configured. The computer is able to detect the
instrument using firmware.
Is there anyone who can help me.
Thanks
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Only one module in LC stack needs LAN card (except Column compartment G1316 which doesn't have LAN card slot), rest of modules are communicating with CAN cables.If LAN-type cable connection, then each module would need to have its own individual IP address.
Try to check in firmware update tool what firmware you have and use one that is compatibille with other modules (Personally I'd update/reload firmware)Hi,
I have problem connecting the G1315B DAD with chemstation. It says the
detector s not propely configured. The computer is able to detect the
instrument using firmware.
Is there anyone who can help me.
Thanks
Told you that I didn't know computers!Only one module in LC stack needs LAN card (except Column compartment G1316 which doesn't have LAN card slot), rest of modules are communicating with CAN cables.