Qualitative Lubricant Analysis Question
Posted: Mon Aug 07, 2006 7:17 pm
I need to be able to distinguish between different lubricants and the method and column that was being used before I got the responsibility for this gives nothing but a very broad non descript chromatogram. I realize that I will probably never be able to a really good separation of the different components within the lubricants but I need some help. The method used was:
column: 1% dexsil 300 on 100/120 mesh Chromosorb WHP packed column
Injection temp: 225 C
Det: (FID) temp 320 C
oven: 155 C hold for 4 min. ramp to 300 C at 12 C/min
I then modified the method to
injection temp: 300 C
Det temp: 340 C
Oven: 90 C hold for 2 minutes ramp 20 C/ min to 200 C then ramp to 320 C at 6 C/min
and got some separation (a broad "hump" with peaks on top) which was good enough for me
BUT I could not reproduce this
the samples were 1% v/v lubricant in hexane and 1 ul injections
So basically I am wondering a couple of things:
1. Is the problem most likely an injection problem (I am doing manual injections)
2. Is it a method problem or column problem
and more generally
3. Is there any rule of thumb for what the injection temperature should be with respect to the boiling points of the analytes or solvents?
I will try to get the chromatograms posted
column: 1% dexsil 300 on 100/120 mesh Chromosorb WHP packed column
Injection temp: 225 C
Det: (FID) temp 320 C
oven: 155 C hold for 4 min. ramp to 300 C at 12 C/min
I then modified the method to
injection temp: 300 C
Det temp: 340 C
Oven: 90 C hold for 2 minutes ramp 20 C/ min to 200 C then ramp to 320 C at 6 C/min
and got some separation (a broad "hump" with peaks on top) which was good enough for me
BUT I could not reproduce this
the samples were 1% v/v lubricant in hexane and 1 ul injections
So basically I am wondering a couple of things:
1. Is the problem most likely an injection problem (I am doing manual injections)
2. Is it a method problem or column problem
and more generally
3. Is there any rule of thumb for what the injection temperature should be with respect to the boiling points of the analytes or solvents?
I will try to get the chromatograms posted