Splitless Injection Issues
Posted: Thu May 11, 2017 8:34 pm
We are running an Agilent 7890A coupled to a 5975C MS.
The system is a few years old, and we normally run in split mode.
We have had the need to increase our sensitivity, and tried doing a splitless injection for the first time. When loading the method in Chemstation for splitless mode the inlet pressure bounces up and down, and obviously the GC never reaches ready status. The system is tight and I can detect no leaks. I tried doing a 1:2 split to see if that would show similar symptoms, and it works fine. We normally run in split mode at 1:150. I am baffled. Short of pulling the EPC unit and sending it out for a refurbish, just wondering if any one might have any other ideas as to what might be the cause. Any input would be greatly appreciated.
The system is a few years old, and we normally run in split mode.
We have had the need to increase our sensitivity, and tried doing a splitless injection for the first time. When loading the method in Chemstation for splitless mode the inlet pressure bounces up and down, and obviously the GC never reaches ready status. The system is tight and I can detect no leaks. I tried doing a 1:2 split to see if that would show similar symptoms, and it works fine. We normally run in split mode at 1:150. I am baffled. Short of pulling the EPC unit and sending it out for a refurbish, just wondering if any one might have any other ideas as to what might be the cause. Any input would be greatly appreciated.