"Host System Not Ready" error with Agilent 7890

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
This GC has both FID and MSD. I used FID detector only. I can still control the GC-FID in ChemStation. But the "Host not ready" is on all the time. There is no data collected for all injections. Any idea? Thanks in advance.
Hello

The best way to sort it out is to configure second instrument with FID only (only GC without MSD). It should do the trick.

Regards

Tomasz Kubowicz