contamination with UPLC-MSMS

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
when working with our UPLC-MSMS-system i sometimes get this problem. a sample injected is very concentrated and after this injection every blank-injection contains the analyte-peaks. so i think the system gets contaminated somewhere. it takes days of injecting blanks before my blank are blank again.
at what point of the system sits the contamination?
does anybody know any easy procedure to get rid of this contaminations?

thanks i.a.

albert

Depends on the sample... if it's a pretty clean sample, I'd look into the injector. If it isn't so clean, I'd check the MSD...
Thanks,
DR
Image

the problem is not related to the matrix. it also hapened if e highconcentration in clean methanol was injected. the problem must be in the UPLC becouse i get a nice peak at the right RT, if the contamination hapened in the MS i would expect a steady background. But where in the LC did it hapen? and what is the easy way to solve the problem?

Since the retention time of the carryover is the same as the sample, it has to be coming from the injector. I have noticed that carryover tends to get worse as the column pressure goes up. I attribute this to fittings and seals flexing slightly and allowing the sample to penetrate into places it would not normally go. I suspect in your case that there is something on the injection valve that is not quite tight enough.

Also, check out the discussion on needle wash.
Mark Tracy
Senior Chemist
Dionex Corp.

^ yep - check the needle wash - is it full? primed? appropriate?
Thanks,
DR
Image

I'm a little late, but let me add a hint anyway.

If your really doing uplc as I understand uplc (means high! pressure) then you have to be sure that the injection part of the autosampler has NO dead volume at all.
If you change for example the rheodyne but don't use a new sample loop you get a dead volume and ANY high conc injection results in carry over problems which may last for up to 100 injections!

To chech if this is your problem run the system with e.g. 100% of acetonitrile and open the fittings of the sample loop so that any contamination / carry over is flushed out. After this procedure the carry over problem should be solved.

Hope this helps
cu/2, Andreas
6 posts Page 1 of 1

Who is online

In total there are 26 users online :: 2 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry