mV settings in electrchemical detection system.

Posted: Fri Aug 04, 2006 2:38 pm
by paraboxa
Can anyone tell me how to select mV settings for electrochemical detection? I'm working on a methyl transferase with a C18 column, isocratic system. Mobile phase is phosphate-acetonitrile-THF at pH6.2. Samples are patient EDTA blood.

Posted: Sat Aug 05, 2006 11:31 pm
by tom jupille
What, exactly, are you trying to detect? And what, exactly, detector are you using (manufacturer, and model number)?

Posted: Tue Aug 08, 2006 9:19 am
by Markus Laeubli, Metrohm
The mV setting (working potential) defines your detector selectivity. An optimized setting may discriminate interfering peaks.
The working potential depends on the component to be analyzed.