Hi Salma
First exuse my bad english
Usualy I recomand to customers to use "adjusted amount" in there reports. "Adjusted amount" take all parameters filled in the sequence rows (sample amount, sample volume, dilution factor...etc) in his calculations. For example, if the sample & standard volume are different, "Amount" give you wrong result but "adjusted amount" give you the right one. Why? Simply because when you use "Amount", TotalChrom calculate the amount using the informatios stocked in the method. When you use "Adjusted amount", the software use the new informations intreduced in the sequence. For this reason, when you need to introduce any new parameter (ex:sample volume) do it in the sequence not in the method. I recomand you also to use automatic calibration in the sequence. This way is very usuful especilly when you have lot of samples & an autosampler.
For the BF, you must know that your pic must have at least 20pts to be integrated propely. I don't know if you remember the factor between "Sampling rate" & "suggested BF". In general it is equal to 20. How to choose the right value of BF? Run your sample with the defaults values of BF (1) & sampling rate (20pts/s). After that report the BF value suggested by the software & divied the sampling rate by this value. This will be your new sampling rate. Take care to choose the next higher sampling rate value available in the software. Ex: BF=6, Calculated sampling rate=3.33, Nearst available softwar's values= 2.5 or 5. Choose 5.
I hope that this modest informations will help you

, don't hesitate to contact me for any further infos.
Regards