-
- Posts: 1
- Joined: Fri Apr 04, 2014 11:29 pm
I read through the user manual and determined that the source of my baseline problems were probably due to cell dirtiness. I only use water and sugars in my HPLC analysis and therefore there might have been residual sugars left in the cell that left a film. Anyways, I attempted the cleaning procedure outlined in the user manual, which was 50 mL acetone , 100 mL H2O, and 50 mL of 0.1M nitric acid followed by extensive purging of my flow lines with H2O (i.e., left overnight). My flow rate was 0.500 mL/min and the pressure was consistent. After refilling the reference cell and balancing, there was a new issue.
Instead of baseline issues, the RID LED screen now displays the message "ALAM", which pertains to the Total Energy value of ~1600, when it should be from 6000-9000. This means that there is not enough light reaching the cell. The lamp has only been run for ~2400 hours and should not have to be replaced until about 20,000 hours. I also increased the Lamp Voltage from 3.89 to 5.00, which is the maximum recommended voltage.
Currently, I am trying another cleaning method using IPA (isopropyl alcohol), Ethanol:Chloroform (80:20), and IPA again followed by extensive flushing with H2O.
My flow rate again is 0.500 mL/min at room temperature. The cell temperature has always been 40 C.
Is there anything else I should try if this does not work, before ordering a new lamp?
