Can anyone tell me how to select mV settings for electrochemical detection? I'm working on a methyl transferase with a C18 column, isocratic system. Mobile phase is phosphate-acetonitrile-THF at pH6.2. Samples are patient EDTA blood.

What, exactly, are you trying to detect? And what, exactly, detector are you using (manufacturer, and model number)?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

The mV setting (working potential) defines your detector selectivity. An optimized setting may discriminate interfering peaks.
The working potential depends on the component to be analyzed.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
3 posts Page 1 of 1

Who is online

In total there are 95 users online :: 0 registered, 0 hidden and 95 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm

Users browsing this forum: No registered users and 95 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry