IPA flushing [August 18, 2004]
Posted: Sun Aug 29, 2004 7:18 pm
by admin
IBy rs on Wednesday, August 18, 2004 - 01:58 pm:
What's IPA flushing of the column
By Ann on Wednesday, August 18, 2004 - 02:18 pm:
Flushing with isopropyl alcohol, also known as 2-propanol or isopropanol.
Hope that helps :o)
Ann
By rs on Wednesday, August 18, 2004 - 04:11 pm:
sure it does...thanks
By Ann on Thursday, August 19, 2004 - 02:46 am:
You're very welcome!
Ann
By syx_gf on Thursday, August 19, 2004 - 05:51 pm:
Is it ok to flush the column with IPA? I always flush the LC system without column using flow rate 2.0 ml/min in 1 hours.
By Mark on Friday, August 20, 2004 - 05:51 am:
to syx_gf,
You certainly can flush columns with IPA, I do it routinely for some natural product work I do, lots of nasties held pretty tightly to my RP column. You will have to watch the back pressure carefully of course since the IPA viscosity is very much greater than the usual LC solvents, you won't be able to run a 2 mL/min flow rate through your column probably.
Mark