My friend told me that GC/MS/NCI is a good detector for halo compound but they said that the stability of signal response is not good, I mean signal vary much in each injection. Is it true?
-
- Posts: 145
- Joined: Sat Sep 25, 2004 3:09 pm
My experience is yes but I don't know why. Probably because of source contamination during sample injection which leads to signal fluctuation. Nonetheless, NCI is less prone to matrix interference for sure. I suggest you to use good I.S. such as isotopically labelled standards for monitoring the changes.
-
- Posts: 9
- Joined: Sat Jul 29, 2006 1:08 pm
I think that it depend also of reagent gaz quality & flow fluctuations...
Reda RADI
PE Instruments Service manager
Algeria
PE Instruments Service manager
Algeria
-
- Posts: 4
- Joined: Sun Jul 30, 2006 4:29 pm
Thanks for your information.
Reda:
I read the data and noticed that the flow of reagent gas (CH4) is quite steady and I guess it is regulated by the GC in certain "psi".
chhubert:
Great to hear that you have same observation. At least I think the problem is not caused by inproper setting. Our situation is that the signal fluctuation also observed just by injecting the standard solution. And I would like to use it for quantitation. So I quite worry about the possibility although many literature papers said it can. Also, do you find the linearity range is very narrow in NCI?
Its very expensive to use isotopes for IS, any cheaper & best alteratives?
Reda:
I read the data and noticed that the flow of reagent gas (CH4) is quite steady and I guess it is regulated by the GC in certain "psi".
chhubert:
Great to hear that you have same observation. At least I think the problem is not caused by inproper setting. Our situation is that the signal fluctuation also observed just by injecting the standard solution. And I would like to use it for quantitation. So I quite worry about the possibility although many literature papers said it can. Also, do you find the linearity range is very narrow in NCI?
Its very expensive to use isotopes for IS, any cheaper & best alteratives?
-
- Posts: 145
- Joined: Sat Sep 25, 2004 3:09 pm
Sky,
I did have similar observation when simply injecting standard solution. However, as you may know that CI source is more prone to contamination, probably due to presence of methane at relative high pressure and low temperature in the ion source, injecting standard solution may still make the source dirty to a certain extent. In my analyses, I also use homologues as I.S. but performance is not as good as isotopically labelled ones. It's just a compromise. Linear range also depends on whether a good I.S. is used, which two orders of magnitude is not a problem.
I did have similar observation when simply injecting standard solution. However, as you may know that CI source is more prone to contamination, probably due to presence of methane at relative high pressure and low temperature in the ion source, injecting standard solution may still make the source dirty to a certain extent. In my analyses, I also use homologues as I.S. but performance is not as good as isotopically labelled ones. It's just a compromise. Linear range also depends on whether a good I.S. is used, which two orders of magnitude is not a problem.
5 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 18 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 18 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science