HPLC-RID for sugar (50 ppm)

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hi,

I'm planning to analyze sugar (glucose, fructose) with the lowest concentration of 50 ppm. My instrument will be HPLC Agilent 1100 (with RID 1100 series also).

Is it possible to analyze sugar with RID for concentration of 50 ppm?
What is the possible column or mobile phase?

Thanks
Hi,

Yes, it will be possible. You can find plenty of methods if you search for "HPLC-RI sugars".
An RID is too insensitive. The best concentration I could detect is 1000 ppm or 0.1% (30 years ago). Your better of to use a Biorad Aminex column (or equivalent) and 210 nm UV.

Check AOAC or ICUMSA (International Commission for Methods of Sugar Analysis) for an appropriate method based on your sample matrix and expected concentration.
Got ELSD ?
We have also been looking into methods for Sugars and Dietary Fiber and they both use RID. I am wondering if LC/MS would be a good option or would there be problems ionizing the sugars?
The past is there to guide us into the future, not to dwell in.
5 posts Page 1 of 1

Who is online

In total there are 18 users online :: 3 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], deevencoo3, Semrush [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry