AA » Sat Aug 12, 2006 2:26 am
The 2695 and the 2795 are two very different instruments. Although the solvent delivery is very similar, the autosampler part couldnt be more different. To put it as simply as possible, in th 2695 the needle is part of the flow path which gives very reproducible (low %RSD's) injections. The 2795 is a sip and spit type, that is to say sample is drawn into the needle and then placed into a loop. If everything is not exactly perfect (weak solvent, strong solvent, and sample diluent matching, loop and needle matching, vial type, cap mat type), poor injection reproducibility is the result. The early units were much worse than the later and current units due to some hardware modifications, firmware changes and added calibration routines. If you bought the unit on the so called "grey market" (not refurbished by Waters), it is likely one of those early units and you would be lucky to get less than 2% RSD. A currently produced 2795 should always give less than 1% RSD for area. After all is said and done a 2695 will ALWAYS out perform and 2795 when it comes to injection reproducibility. The 2795 has a lower system volume, better flexability as it can sample from vials or 96/384 well plates. Carryover is often better on the 2795 if you choose the correct wash solvents and wash volumes. As well, injection to injection cycle time is much better on the 2795. Solvent delivery charateristics from both the 2796 and 2695 are identical.