Two in progress status when running GC Agilent

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Hi all,

Please help me out, I am running GC Agilent using Chemstation sofware with 1 sequence. The sofware recently apprear two status of in progress, please see the attached picture.

https://goo.gl/photos/6TEaWhCTcAdDrL6u8

Thank you guys so much for your help.
Go to that datafile location in Windows explorer. You will find it has an extra file that your other datafiles don't have (acquiring.txt i believe?) Once you remove that, in progress should go away.
Go to that datafile location in Windows explorer. You will find it has an extra file that your other datafiles don't have (acquiring.txt i believe?) Once you remove that, in progress should go away.
Hi,

I did as your support and it was actually fixed。 But I still wonder why this extra file appeared, because I did the same actions with other location files (creating sequence, choose data location,run sequence)。Others appear normal, but this one is quite different.

Thank you so much for your support me!
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry