I am setting up a headspace SPME procedure to measure Ethyl Carbamate/Urethane in alcoholic beverages. I am following the exact same protocol as another lab as far as the SPME fiber, GC oven program etc. But I am not having any luck with my analysis. The GC-MS is working fine. The only difference is the column- I am using a HP-5MS whereas the other lab used a DB-WAX column. Simce, EC is a relatively polar compound, could this be the reason for me not getting any EC peaks even at 1 mg/L conc?
Thanks!

It could be. SPME depends on focussing the peaks at the top of the column as they elute from the fibre. On a non-ploar column the ethyl carbamate is less strongly retained, and so less focussed than on a PEG column. If you look really closely you migh find a very low broad peak for the ethyl carbamate.

Why struggle ? - use a PEG column that you know will work.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry