Linearity of calibration curve

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Dear all,
How should one define the linearity of the calibration curve obtained?
If R2 > 0.99 is not sufficient to claim that our calibration curve is linear, what else do I need....standard deviation of the slope? residual plot? or anything? Please help to explain how someone can claim that the calibration curve is linear, especially during the submission of paper for the journal.

also would appreciate if anyone can help to explain the real meaning and the implication of these parameter (standard deviation of the slope, residual plot and so on) to the calibration curve

Thanks
Kenny

For any given correlation coefficient the points might cluster around a straight line or a curve. To demonstrate linearity over the range of interest you must plot residuals against concentration, ie. the distance from the points to the regression line, parallel to the detector response axis. You need at least 10 points for the plot, some say 8 but that's pushing it. Having got your plot there are a couple of possibilities. Either the residuals are normally distributed around the zero line in which case the calibration function is linear (but not necessarily with a value of r close to 1). On the other hand the residuals might seem to form a curve, in which case you should repeat the calibration to check whether you get a similar curve. If you do then you can do one of two things. You can either fit your calibration function to a polynomial, or you can declare that the calibration is linear within certain criteria that you must decide, based on some context (data quality objective, SOP requirement etc) that only you can know about. I suppose you could use the same criteria for the polynomial fit. It's important to realise that there is no single number answer to the question – is this calibration linear ?

The standard error of the slope is only significant for a linear function and then only when you want to estimate uncertainty due to lack of fit.

I may have oversimplified this a bit, but then I am not a metrologist, I am a normal person !
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry