I'm trying to figure out the answer to this question. Our Agilent instruments use make-up gas with their FIDs, but our Perkin Elmer Autosystem XL does not. Why? I can't seem to find this answer anywhere... We use a hydrogen flow of 45 mL/min and air at 450 mL/min, versus 40 mL/min and 400 mL/min on the Agilents. Carrier pressure (He) 9.5 psig on the PE versus 35 psi on the Agilents. Does it have something to do with the differences in gas flow? Sorry, I'm a newbie...
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Welcome to the forum.
Make-up gas was originally used to flush the eluent from capillary columns through the large volumes of plumbing in FIDs the were made for packed columns. As companies redesigned their FIDs to work with the lower carrier gas flows of capillary columns some of them dropped the make-up gas - the fuel hydrogen is enough to flush the eluent quickly through the flame.
However, nitrogen as make-up gas also increases signal:noise by slowing diffusion at the base of the flame where analyte is hydrogenated by hydrogen radicals diffusing upstream form the combustion zone. How big the improvement is varies from a few percent to 30% in different studies - some companies consider this worth the extra plumbing and cost, others not.
Peter
Make-up gas was originally used to flush the eluent from capillary columns through the large volumes of plumbing in FIDs the were made for packed columns. As companies redesigned their FIDs to work with the lower carrier gas flows of capillary columns some of them dropped the make-up gas - the fuel hydrogen is enough to flush the eluent quickly through the flame.
However, nitrogen as make-up gas also increases signal:noise by slowing diffusion at the base of the flame where analyte is hydrogenated by hydrogen radicals diffusing upstream form the combustion zone. How big the improvement is varies from a few percent to 30% in different studies - some companies consider this worth the extra plumbing and cost, others not.
Peter
Peter Apps
-
- Posts: 2
- Joined: Thu Apr 13, 2017 5:35 pm
Thank you, Peter.
3 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 5 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], deevencoo9, Google [Bot], Majestic-12 [Bot], Semrush [Bot] and 21 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], deevencoo9, Google [Bot], Majestic-12 [Bot], Semrush [Bot] and 21 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science