by
KM-USA » Sat Mar 25, 2017 1:41 pm
Hi...i am a new user here. I am trying to get Chemstation ver A10.. to communicate with my 5890 series II GC after upgrading the computer to Windows XP. We have a PCI GPIB card in the computer but only an HPIB card in the GC. Agilent thought this should work. In the configuration, there is only a option to put an address in for a GPIB card. The HPIB address doesn't seem to work. Anybody out there using this version of chemstation with the 5890? Do you need a GPIB card for the GC or did you get it to communicate with the HPIB.
HPIB and GPIB are essentially the same thing, I think HPIB was just Hewlett Packards proprietary name for them back in the day.
Most have a set of switches on them that you use to set the address physically on the card, then you have to put that number into the software. I would think A10 should still support the HPIB/GPIB interface, but if you also have the 9pin serial port on the computer and instrument you can configure communications that way. I did it once when the HPIB card went out in the computer. Should be instructions in the manuals for that as well.
Hatty, the GPIB cable plugs into the GC, right? Then the GC has the GPIB board inside it, good. You'll also need the 9-pin remote cable from the GC plugged into the computer.
The computer itself needs the PCI GPIB card configured. That board should be either an 82341 or an 82350. This can be tricky, as when configuring this through the A.10, you must enter as hp82341 (lower case only) even if the card is actually 82350 (don't ask me why); you must also enter the bus address as 30 (again, don't ask me why).
In Configuration Editor, you must also enter the address for the 5890 itself; most likely this would be the factory address, but you can confirm module address using the front panel. An autosampler would have its own address, and also must be connected with GPIB and remote cable to the control box.
If the system worked before the computer change/death, and you transferred the GPIB board from the old computer to the Windows 7 computer, this should work if configured correctly.