Hi,

We have an old PE LC-95 UV detector that is working fine at wavelength of 313 nm. Currently, we are going to use it at 248 nm for a HPLC analysis. However when we changed wavelength, it can not be zeroed. It always showed "-1" under 285 nm when push the A/Z button (ok above 285 nm). We have tried washing of flow cell with string acid but still did not work. Anybody has clue? Thanks.

:?:

Dear lu_pharm:

I had used old P-E detectors (not exact model). Perhaps I can offer some tips:

1. Check if it is a variable (wavelength) detector.
2. When changing wavelength, you should go from the ‘low side.’ For example, you should set it at 280, and then increase gradually to 285nm.
3. Passing strong acid may not help. You should run mobile phase through it long enough (may be overnight?).
4. A value of –1 is not really a bad thing. Just inject your standard, to see if you can get a good peak, and if the baseline is acceptable.

Good luck.

Alfred
2 posts Page 1 of 1

Who is online

In total there are 34 users online :: 2 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry