Wavelength Rnge for PDA Peak Purity

Posted: Mon Jul 10, 2006 5:02 pm
by gbalock
Typically when we perform a method validaton, we evaluate the peak purity from about 210nm to 300nm. Is this enough or is it too much. What wavelength range do you use, or do you use a range around the analysis wavelength?

Posted: Tue Jul 11, 2006 12:15 pm
by amitavaghosh
we should use wavelength range 210 to 400nm

It depends...

Posted: Tue Jul 11, 2006 5:19 pm
by juddc
Your wavelength range for peak purity really depends upon two factors: The UV absorbance spectrum of your compound and the background absorbance of your mobile phase. If you are assaying a sunscreen with an absorbance maximum of 360 nm and you're using an acetate based mobile phase, 210-300nm won't do too well. You'd want to be between 230 - 450 nm. On the other hand, if you have an abs max at 234 with nothing beyond 250 nm and are using a phosphate-ACN MP, you might want to look at 205-270. It all depends on what you'relooking at and the conditions you're using.

How mus=ch is enough?

Posted: Mon Jul 17, 2006 9:20 pm
by gbalock
Juddc,
Thanks for the input. To rephrease the question how much is enough, 50nm on either side of yoiur analysis wavelength, 100nm?