Interfering peaks in the LCMSMS technique
Posted: Mon Jul 10, 2006 7:22 am
Dear friends,
We have been working on drug analysis in human biological matrices (plasma, serum, blood, urine) during these last few years using several API 4000 or Waters Premier. Sometimes for unexplained reasons we face interfering peaks in the LCMSMS technique at the retention time and m/z of parent and daughter that are very big (more than 50% of LLOQ). We usually check for the source of this interference, by cleaning the instrument, injecting mobile phase, or injecting the solvents used in extraction (by evaporating the amount usually used and then reconstitution and injecting under the same conditions). Also when this happens we change all the devices and tools, e.g., autosampler by using a manual injector, or even the whole HPLC, micropipettes, etc. Unfortunately these procedures sometimes do not solve the problem.
My question is, if these interferences are not contaminations of the analytes of interest, how come we get a quantifiable signal (more than 50% of LLOQ) at the same retention time, parent and daughter m/z? This is against the concepts of selectivity and specificity of the LCMSMS technique. Is there any explanation of interference in LCMSMS technique that is "opposite" of ion suppression? Would I be expecting extra signals due to plasma components that, although does not have the same m/z, yet still rise up as a quantifiable peak?

We have been working on drug analysis in human biological matrices (plasma, serum, blood, urine) during these last few years using several API 4000 or Waters Premier. Sometimes for unexplained reasons we face interfering peaks in the LCMSMS technique at the retention time and m/z of parent and daughter that are very big (more than 50% of LLOQ). We usually check for the source of this interference, by cleaning the instrument, injecting mobile phase, or injecting the solvents used in extraction (by evaporating the amount usually used and then reconstitution and injecting under the same conditions). Also when this happens we change all the devices and tools, e.g., autosampler by using a manual injector, or even the whole HPLC, micropipettes, etc. Unfortunately these procedures sometimes do not solve the problem.
My question is, if these interferences are not contaminations of the analytes of interest, how come we get a quantifiable signal (more than 50% of LLOQ) at the same retention time, parent and daughter m/z? This is against the concepts of selectivity and specificity of the LCMSMS technique. Is there any explanation of interference in LCMSMS technique that is "opposite" of ion suppression? Would I be expecting extra signals due to plasma components that, although does not have the same m/z, yet still rise up as a quantifiable peak?